Pacemaker Nomenclature

Discussion in 'ITE Keywords' started by usnavdoc, 11.26.08.

  1. usnavdoc

    usnavdoc Senior Member

    Joined:
    04.05.05
    Messages:
    917
    Status:
    Attending Physician
    SDN 7+ Year Member

    SDN Members don't see this ad. (About Ads)
    PACEMAKER NOMENCLATURE –
    [FONT=Arial,Bold]
    NASPE/BPEG Revised in 2002 NBG Pacemaker Code
    Position I Position II Position III Position IV Position V​
    .
    (Chamber Paced) (Chamber Sensed) (Response to Sensed Event) (Programmability, (Multisite Pacing)
    O = none O = none O = none Rate Modulation) O = none
    A = atrium A = atrium I = inhibited A = atrium
    V = ventricle V = ventricle T = triggered O = none V = ventricle​
    D = dual (A + V) D = dual (A + V) D = dual (T + I) R = rate moduation D = dual (A + V)



    The first two positions of this code (Chamber Paced and Sensed) are relatively
    straightforward. The third position is the most confusing. The most frequently used programs are
    the DDD (dual chamber pacing and sensing, both triggered and inhibited mode), VVI (for single
    chamber, ventricular pacing in the inhibited mode), VDD (ventricular pacing with atrial tracking),
    and DDI (dual chamber pacing and sensing, but inhibited mode only).

    The third position is described as follows:
    D - (dual): In DDD pacemakers, atrial pacing is in the inhibited mode (the pacing device
    will emit an atrial pulse if the atrium does not contract). In DDD and VDD pacemakers,
    once an atrial event has occurred (whether paced or native) the device will ensure that an
    ventricular event follows.

    I - (inhibited): The device will pulse to the appropriate chamber unless it detects intrinsic
    electrical activity. In the DDI program, AV synchrony is provided only when the atrial
    chamber is paced. If on the other hand if intrinsic atrial activity is present, then no AV
    synchrony is provided by the pacemaker.

    T - (triggered): Triggered mode is only used when the device is being tested. The pacing
    device will emit a pulse only in response to a sensed event.

    The VDD pacemaker is used for AV nodal dysfunction but intact and appropriate sinus
    node behavior. DDI is rarely used as the primary mode of pacing. The DDI pacer is used for a
    patient with a dual-chamber pacemaker that has episodes of paroxysmal atrial fibrillation. DDI
    prevents high ventricular rates. Some DDD pacemakers are programmed to enter the DDI mode
    when high atrial rates occur.

    The fourth position, rate modulation, increases the patient's heart rate in response to
    "patient exercise". A number of mechanisms (vibration, respiration, and pressure) are used to
    detect "patient exercise". As the exercise wanes, the sensor indicated rate returns to the
    programmed mode.

    The fifth position describes multisite pacing functionality. Atrial multisite pacing is being
    investigated as way to prevent atrial fibrillation. Ventricular multisite pacing is a treatment for
    pacing a patient with dilated cardiomyopathy.

    There are four types of pacemakers: asynchronous, single-chamber synchronous,
    double-chamber AV sequential, and programmable.

    • Asynchronous or fixed-rate (AOO, VOO, DOO) – discharge at a preset rate that is
    independent of the inherent heart rate.

    • Single-chamber synchronous or demand (AAI, VVI) – discharge at a preset rate only
    when the spontaneous heart rate drops below the preset rate.

    • Dual-chamber AV sequential pacing (VDD, DVI, DDD) – usually uses two electrodes, one
    in the atrial appendage and one in the right ventricular apex. The atrium is stimulated to
    contract first, then after an adjustable PR interval, the ventricle is stimulated.

    • Programmable pacemakers – pacing rate, pulse duration, voltage output, and R-wave
    sensitivity are the most common programmable functions.
     

    Attached Files:

    Last edited: 11.30.08
  2. usnavdoc

    usnavdoc Senior Member

    Joined:
    04.05.05
    Messages:
    917
    Status:
    Attending Physician
    SDN 7+ Year Member
    My chart didnt work out so well so I included a word file. One type of question I have come across related to this is that they give you an EKG tracing and ask you to determine which type of pacemaker the patient has. I have also included that in the word file as I cant seem to get it to paste in the thread.
     
  3. Freibi

    Freibi

    Joined:
    11.13.06
    Messages:
    132
    Location:
    Black Forest
    Status:
    Fellow [Any Field]
    SDN 2+ Year Member
    Great overview! Thank you! Did you mean (in the last sentence above) "... the device will ensure that a ventricular event follows"?

    For me, the third position was always hardest to understand, until I realized that it describes the reaction of the pacer to the heart, not the other way round. I.e., you have to ask: does the pacemaker get triggered or inhibited or both by what the heart is doing?
     
  4. usnavdoc

    usnavdoc Senior Member

    Joined:
    04.05.05
    Messages:
    917
    Status:
    Attending Physician
    SDN 7+ Year Member
    Yeah thanks for the corection. I always had trouble with the third position as well.
     