Coke vs Pepsi. Both are marquee products from their respective manufacturers. But agree with medgator you want your hardware and software to be maximally compatible.
This thread is old but still relevant, so I will give my opinion at the risk of being considered spam according to the forum rules. In the last years it seems that Elekta spent so much money and effort in developing their MR-linac, that forgot to finish and improve other products like Versa HD or MOSAIQ. Here it goes a comparison of several points from my view:
Truebeam+ARIA is much more user friendly than VersaHD+Mosaiq, although any ot them is probably better than a mixed-vendor combination for compatibility reasons. This has an important impact in the learning curve of new staff (mainly therapists and probably physicists, not so much for physicians).
In Varian, the linac control is more compact: with Elekta the therapists need to use several computers and a mix of very different pieces of software to control the linac and the different image systems. For example, the registration of the CBCT images with the planning CT is done in the XVI program, but if you use orthogonal kV images the registration with the DRRs is done in Mosaiq with a completely different interface. At the end, people tend to forget the less used one and use CBCT for everything, even for cases where kV-kV may be appropriate.
Also, the Elekta 6DoF couch (Hexapod) needs its own software and its own computer, with its own keyboard etc, and the same happens with the Elekta add-on microMLC. Hexapod also needs to attach a bridge to the couch with IR reflectors. The Varian 6DoF couch does not need this and is more integrated in the system.
Varian MV and kV panels are mounted in robotic arms and can be moved remotely from the control room, which is not possible in Versa HD. In Elekta it is even neccesary to change the collimator and filters of the kV tube manually. In Varian the source-EPID distance is variable and the panel can go up to the isocenter (which is useful for linac QA), while in Elekta the distance is fixed. I think Varian recently introduced tools to automate the linac daily QA, while Elekta doesn’t have anything similar.
The Elekta couch needs to attach a head extension for most treatments, and the area of the couch where the extension board is attached produces very high attenuation (>20% for some angles), limiting the usable entry angles in some treatments. Varian linacs don’t have this issue.
The mechanical accuracy (gantry sag, etc) is better in Truebeam. The difference is not very relevant for conventional treatments but it it probably importannt for SRS.
And ARIA is more intuitive than MOSAIQ, in my opinion. Sometimes even the Elekta support specialists do not understand Mosaiq. They have added many tools and options over the time, which makes it more complex, but still keep the same old structure and design.
[Edit: and I forgot about it... I have been told that in "myvarian" you can find useful things. In "Elekta community" appart from the pdf manuals of the different products there aren't anything useful]
So, I wouldn’t say it is like Coke vs Pepsi.