Hello all,

Had a patient referred s/p L5-S1 ALIF and subsequent SCS placement with a paddle lead up to T9. The referring doc is asking for MBBs at L4 and L5 DR with subsequent possible RFA. Anyone have experience with doing an RF with a stim in place? Thanks!
 

lobelsteve

SDN Lifetime Donor
Lifetime Donor
10+ Year Member
May 30, 2005
13,730
3,602
Canton GA
www.stevenlobel.com
Status
Attending Physician
No issues. Place ground pad near RF location on skin.
 

specepic

10+ Year Member
Feb 2, 2008
1,759
252
New England
Status
Attending Physician
The official word from BS is to not do it (RFA). Until the day of tort reform, I do not therefore. I can imagine that it prob wouldnt be an issue in reality
 

emd123

7+ Year Member
Feb 25, 2010
4,206
1,475
Status
Attending Physician
The official word from BS is to not do it (RFA). Until the day of tort reform, I do not therefore. I can imagine that it prob wouldnt be an issue in reality
I believe the concern is that it could hurt the battery and stim equipment, not the patient. The reps have told me the same thing, just turn the stim off. I've never seen it be an issue, it's so low voltage and current to begin with, and you're putting that same current up to their spine normally anyways.
 

clubdeac

10+ Year Member
Aug 16, 2007
3,376
605
Status
Attending Physician
I had one patient come back afterward stating the RF drained the IPG. He said after the procedure he was having to charge the batter daily whereas before it was more like weekly. Didn't know what to make of it other than replace the battery
 

bedrock

Member
10+ Year Member
Oct 23, 2005
4,160
1,095
Status
Attending Physician
I've done about 10-12 RF on patients with SCS. Just have them turn off SCS, put ground very close to RF cannulae, no issues.
 

lobelsteve

SDN Lifetime Donor
Lifetime Donor
10+ Year Member
May 30, 2005
13,730
3,602
Canton GA
www.stevenlobel.com
Status
Attending Physician
Did left side today, right side 2 weeks ago. IPG on right flank. Ground pad between IPG and RFA probes.
 

emd123

7+ Year Member
Feb 25, 2010
4,206
1,475
Status
Attending Physician
I've continued without problem. Did RFa on a bladder Stim patient today also. Per urologist and rep, same rules, turn battery off and keep needles and grounding pad far away from battery and leads.
 

olafa

10+ Year Member
Jul 26, 2007
653
30
Status
Attending Physician
I had one patient come back afterward stating the RF drained the IPG. He said after the procedure he was having to charge the batter daily whereas before it was more like weekly. Didn't know what to make of it other than replace the battery
that does not make sense? How would that happen it IPG is off. Have done it without problems, SCS turned of during procedure and not a prob.
 

clubdeac

10+ Year Member
Aug 16, 2007
3,376
605
Status
Attending Physician
that does not make sense? How would that happen it IPG is off. Have done it without problems, SCS turned of during procedure and not a prob.
Yeah I shrugged my shoulders and said the same thing. That's what he told me though
 

ctts

7+ Year Member
Oct 15, 2010
85
2
Status
If you want to be safe, can do a true bipolar RF which does not require grounding pad. I have done this on one patient. Twice as many needle sticks for the patient though. And no additional reimbursement for the additional work.