Most of the time, the lumbar entry works just fine, especially in patients with limited thoracic and cervical mobility. The caveat is that if you meet significant obstruction on advancement in the thoracic or cervical spine, switch to a lateral or contralateral view frequently to assure the tip of the lead is not diving anteriorly (There are reports of the lead diving into the cord itself and then advancing cephalad within the cord).
HOWEVER, for patients who are active and mobile, there is significant lead cephalad-caudad movement due to thoracic spine flexion, resulting in erratic coverage of upper extremity and cervical spine pain or even overt shocking of the patient. This is due to the anchor point being 30-40cm away from the lead tip with the lead extending over a spinal segment that may have significant flexion/extension, causing distal migration of the lead tip in thoracic flexion. On occasion, I have seen distal migration of the leads requiring surgical revision (not easy given the fibrosis bands that are then present above the tip of the migrated leads).
Early on after implantation, there may be a greater tendency towards lateral migration of the lead tips especially if there is cephalocaudad movement of the lead during thoracic flexion since the bands of fibrosis tissue that forms over the posterior aspect of the leads may be delayed, or the linear channels in this fibrosis tissue may become elliptical shaped allowing slight lateral motion. This can partially be eliminated by crossing the lead several times back and forth across the midline during advancement through the plica medialis dorsalis, but if there are two leads advanced in this manner, there will be obligatory stacking of leads, and if they cross in a segment of the spine that has limited AP diameter, the cord may be impacted.
Finally, during such extended advancement of the leads, there is an inherently greater resistance encountered to advancement of the leads, sometimes causing buckling of the leads into the lateral gutter of the epidural space. If this is not corrected at the time of anchoring (by retraction of the lead until the lateral bowing is corrected), the patient may experience radicular pain at the area of lateral buckling of the lead. Routine use of a lead introducer can alleviate much of this (needle-lead blank/guidewire-remove needle-introducer advanced to full length into the epidural space-removal of lead blank/guidewire-advancement of the lead). One sign of migration of the cervical tip distally may be thoracic or lumbar radicular pain onset.