Thanks for posting this and the article!
What beam angles for IMRT? Aren't you all doing VMAT planning for IMRT?
Dose constraints met - this is on scorecard
PTV coverage - this is on scorecard. What are you looking for on axials? Asking for myself, because I look as well, but if my goal is met, I'm not sure what I'd reject, but I do this not knowing exactly why I do it. EDIT: article delineates this nicely.
Hot spots outside of tolerance - this is on scorecard
105% in breast - on scorecard
MU number - hmm, I guess I should look at that. Not sure exactly what to look for? Elaborate?
SBRT CI - scorecard
Axials first then scorecard - why exactly, Seper, other than dogma or routine? I don't understand. EDIT: from article, I see explanation and pasted below.
1 cm from rectum/50% - contour rectum, remove all but 1 cm - put on scorecard
SpaceOar - yes sir!
Efficiency - yes, this is important. More about initial instruction to dosi, rather than at plan approval? I sort of know what should be AP-PA and what should be multi-field. All the things you're saying should be directed pre-plan, I think?
Arc going through flabby pannus - good point. Although, should be considered pre-plan, right?
...
Useful article!!
To your points, from article:
For 3-dimensional (3D) plans, it is important to ensure that the fields are entering the body at angles that avoid entry through excess normal tissue. In addition, beam shaping with multileaf collimators (MLCs) or other devices should be appropriate for a given target and surrounding OARs. This can be evaluated by directly visualizing each field using the beam’s eye view and is also based on 3D isodose lines overlaid on the computed tomography (CT) images. When treating an area in the neck or thorax, for example, one should ensure the beams are not entering through the shoulders/arms or exiting the oral cavity unnecessarily. For intensity-modulated radiation therapy (IMRT) plans, one should consider the number of fields and their point of entry through the body and fluence patterns. Assessing the field arrangement and collimation may subsequently become important if target volume coverage or OAR dose limits are not optimal and may be improved with additional fields or different beam entry angles.
The number of fields or arcs is also a key factor in the treatment time. A patient undergoing a palliative radiation treatment may not be able to lie on the treatment table for long periods, and a faster treatment may be preferable. Radiation oncologists should also consider that patient mobilization and internal organ motion are increased with longer treatment times.
The DVH must be used with caution. The DVH cannot assess the appropriateness of the targets and OARs. The DVH could report 100% coverage of the PTV by the prescription dose, but the PTV could be delineated incorrectly. Alternatively, 95% PTV coverage may not be met, but there may be a compromise between PTV coverage and OAR constraints, with an accepted sacrifice in PTV coverage to avoid unacceptable toxicity to a surrounding critical OAR. Furthermore, there may be excessive dose spillage through structures not reported within the DVH. Because this information cannot be obtained from the DVH alone, we recommend evaluating the 3D graphical plan qualitatively before proceeding to the DVH.