Hi danielmd06,
The mismatch principle with CT perfusion is still CBV and CBF, but some people use TTP (time to peak) instead of CBF. In any case this is still theoretrical as far as CT imaging is concerned since we have no randomized thrombolytic trial using CT criteria (published or presented so far). What further confuses the picture is that most people use 'visual' inspection to look for mismatch which is highly dependent on the manufacturer of the CT scanner. There is still no standardization of the values of CBF/V and TTP on the CT scanners (we tend to look at the colored images) among the manufacturers (ie GE, Siemens, Philips, Toshiba...). In other words we just visually inspect the colored images to look for so called mismatch, but there is no standardization of the color coding yet. MRIs are still better. We have some trials looking at MRI mismatch criteria with IV thrombolytics in extended time windows, but again the mismatch theory is promising as seen with results from DEDAS, DEFUSE, EPITHET and DIAS2, but doesn't necessarily translate into outcomes. So we still need to further fine tune the mismatch criteria before using it clinically as a definitive marker of tissue at risk. MRI is further along the authentication road than CT perfusion, but a lot of work is still needed. We need to differentiate 'benign oligemia' (tissue that may be slightly ischemic, but will not infarct) from actual tissue at risk. It looks like CT tends to overestimate the tissue at risk. Further trials (perhaps MRRESCUE, and DIAS3/4) could shed more light on this.