Agree with the above.
On the fluoro access, drop some lidocaine on the skin to be kind. You will want to get an en face view of the side port (
Complications Associated with Intrathecal Drug Delivery Systems) which can mean obliquing to some weird angles.
Realize the pump is generally attached on an unstable muscle plane that moves with belly breathing, so keep the patient's relaxed or have them hold their breath at times.
The C-arm generally has a hard time with seeing the 24g Huber needle against the dense pump so drop another needle on the field so you can see exactly where you're poking.
When you think you're in the side port, get a shot across the pump and try to zoom in on the port as sometimes you aren't really in. You should be able to see your needle coming in with good fluoroscopy. I don't have a good online fluoro shot for that but imagine something in that right port.
That's only important if you can't aspirate, as if you get CSF, you know you're in.
You almost never want to inject if you can't aspirate.
Please remember to give a priming bolus after the study if you are able to get CSF as you don't want the patient to go into withdrawal. If you can't get CSF or it's not intrathecal, obviously not important.
The Ascenda catheter is easy to see once contrast is in it, and the tip does have a metal marker that you'll see on a lateral fluoroscopy easily but not as much in AP every time until you've done a few.