R Shoulder Arthroscopy in patient with Inspire Device

Started by DocVapor
This forum made possible through the generous support of SDN members, donors, and sponsors. Thank you.

DocVapor

Full Member
Lifetime Donor
10+ Year Member
Advertisement - Members don't see this ad
I've got the above case coming up. It will be my first time dealing with the inspire device perioperatively at all. The manufacturer website is unhelpful on what to expect and the only article I've found is this one, which isn't super helpful. Patient's generator is located in the right subclavicular region.

We generally do these cases with an ISB for POPC, which after reviewing the R shoulder x-ray I should be able to complete just fine. This surgeon is -meticulous- (if you catch my drift), so the patient will get intubated, but he does them in the lateral position at least.

Obviously I'll leave the device off and hopefully the patient will bring the remote as instructed. Has anyone seen any complications from electrocautery? Damage to the device? Any decreased effectiveness post-op if my block manages to bag the phrenic nerve on that side?

Anything else I'm missing?

Thanks in advance.
 
To draw parallels to SCS, make sure device is off and that you had discussed potential electrosurgical interference or device damage is always possible though not likely. Document both of these. Do block, restart device in PACU. Done.
 
It's a hypoglossal nerve stim that is turned on by the patient when they sleep, the device itself is not "on" all the time. No reason to suspect bovie will interfere / harm it. Doubt you will encounter the lead while doing an ISB. The device has nothing to do with the phrenic nerve.
 
Advertisement - Members don't see this ad
It's a hypoglossal nerve stim that is turned on by the patient when they sleep, the device itself is not "on" all the time. No reason to suspect bovie will interfere / harm it. Doubt you will encounter the lead while doing an ISB. The device has nothing to do with the phrenic nerve.
Yes, I was aware of the patient on/off, just trying to cover my bases here. I admittedly haven't kept up with the devices as I haven't been seeing them, but when I talked with an ENT a few years ago he explained to me they had to implant sensors to detect the diaphragm contraction, and my worry was if there was phrenic weakness due to my block afterwards, would the device pick it up.

I now see the newest units have integrated respiratory sensors so it shouldn't be an issue.

Thanks for the reassurance from both of you.
 
Could always do a supraclav instead, and stay out of the neck altogether. Years ago, Blade posted some article showing equivalent POPC outcomes compared to ISB. It's what I do unless I'm dumb enough to try doing the case under block+MAC.
 
To draw parallels to SCS, make sure device is off and that you had discussed potential electrosurgical interference or device damage is always possible though not likely. Document both of these. Do block, restart device in PACU. Done.
I don’t understand how it’s somehow incumbent upon us to assume responsibility for something the surgeon is doing. I get us being involved in periop PPM/ICD management since it has serious hemodynamic consequences. But SCS, inspire, etc. that’s between the patient and the surgeon. I’m not personally doing anything that’s going to be a problem for their device.
 
I just did ACE questions a few weeks ago and there's a new block the costoclavicular block (CCB) that was noninferior to ISB in arthroscopic shoulder surgery and essentially eliminated phrenic nerve palsy. It seems fairly easy to do, might be worth looking into it.
 
I don’t understand how it’s somehow incumbent upon us to assume responsibility for something the surgeon is doing. I get us being involved in periop PPM/ICD management since it has serious hemodynamic consequences. But SCS, inspire, etc. that’s between the patient and the surgeon. I’m not personally doing anything that’s going to be a problem for their device.

Senior partner in large group was called to a closed door meeting for some bad outcome. On his recounting, the one line that stuck with him was an administrator stating

"Doctor, you assumed full liability the moment you accepted to proceed with the case"
 
Senior partner in large group was called to a closed door meeting for some bad outcome. On his recounting, the one line that stuck with him was an administrator stating

"Doctor, you assumed full liability the moment you accepted to proceed with the case"


The appropriate response is for the large group not to proceed with any high risk cases.
 
1769455903134.png
 


1. Make sure the device is turned off prior to surgery

2. Intubate the patient (ETT) in the OR (device off)

3. Consider turning on the device in the PACU to assist with the Sleep Apnea until the patient is fully awake

4. The ISB on the Right side may interfere with phrenic nerve function affecting the device's ability to sense the movement of the diaghram; this could significantly increase the risk of severe Sleep Apnea postoperatively for the next 24-30 hours. (single shot ISB).

I would request that the patient bring the remote control with him/her. If the shoulder surgery was on the right side then an ISB may be contraindicated for outpatient surgery.
 
I have done costoclavicular block for 1 patient getting shoulder surgery. Patient was a pulmonary cripple that I couldn’t risk phrenic nerve paralysis. It was only one patient but it wasn’t that great for shoulder coverage. Patient needed narcotics in comparison to isb which they get none. Your experience may vary.
 
Advertisement - Members don't see this ad
I have done costoclavicular block for 1 patient getting shoulder surgery. Patient was a pulmonary cripple that I couldn’t risk phrenic nerve paralysis. It was only one patient but it wasn’t that great for shoulder coverage. Patient needed narcotics in comparison to isb which they get none. Your experience may vary.
I agree that interscalene block works best for shoulder surgery, the other blocks are all inferior imho.
 


1. Make sure the device is turned off prior to surgery

2. Intubate the patient (ETT) in the OR (device off)

3. Consider turning on the device in the PACU to assist with the Sleep Apnea until the patient is fully awake

4. The ISB on the Right side may interfere with phrenic nerve function affecting the device's ability to sense the movement of the diaghram; this could significantly increase the risk of severe Sleep Apnea postoperatively for the next 24-30 hours. (single shot ISB).

I would request that the patient bring the remote control with him/her. If the shoulder surgery was on the right side then an ISB may be contraindicated for outpatient surgery.
This.
 
Is that truly possible?
I've had one after a successful Tap block.

We present a case of transient femoral nerve palsy occurring after a TAP block with involvement of the sacral plexus for a patient who had undergone a caesarean section.
 
Have y'all done some shoulder articular branch blocks for these? I enjoy doing them as ultrasound guided blocks in the office with 1 mL of bupi 0.5% as a diagnostic prior to RFA. They won't cover the the skin incision, but 2-4 mL on the lateral pectoral, suprascapular, and axillary nerve's articular branches should get good post-operative analgesia without the phrenic issues.