No. (edit: maybe?)
The Bragg peak is just a feature of the dose/depth curve for protons statistically. It does not exist for a single proton. Of course, it's location depends on beam energy.
As mentioned above and in many places, the far side of the Bragg peak seems to be a region where a lot of biology goes on (manifested simply as highly variable and uncertain RBEs in this region). There is a region of uncertainty for any constituent proton beam.
Any pure proton plan is presumably going to achieve coverage through can aggregate of beams of protons, which have dosimetry that includes a bragg peak within the PTV.
Now, are you telling us that the plans with IMPT include beams where the Bragg peak exists outside of the PTV? This would seem weird and to not be taking advantage of what protons bring to the table.
Photon plans are made up of beams where Dmax is outside of the PTV almost always, but this not an area of dosimetric uncertainty and we can have confidence in the aggregate dosimetry of an essentially infinite beam number.
My guess is that the solutions that are come upon with IMPT are uniformly made up of beams whose Bragg peak is within the planning target volume. (Mathematically speaking, it would be like saying that the
basis set for planning, includes only beams with a Bragg peak within the target volume).
It seems that a hybrid photon/proton plan would be the best at that point.
This makes sense. One could come up with a solution regarding the proton portion of the plan where only a portion of the PTV is covered by protons, and the region of dosimetric uncertainty is within say the GTV. This will be a "proton boost" of uncertain dose but potentially with clear dosimetric advantage regarding nearby OARs (not eating up big 30 Gy volumes in healthy liver or a previously irradiated brainstem for instance).
Probably 4 centers nationally doing this level stuff makes sense. I would not advocate for a boost like this in the overwhelming majority of CNS tumors, prostates, lungs, head and necks or other.
edit: how many gantry positions do you actually use for IMPT? I am not aware of there being anything like RA or even a 6-7 beam IMRT arrangement for IMPT. Usually 2, potentially 3-4 gantry positions? Factoring in range uncertainty as well, it is just not clear to me how you can ensure that the highly uncertain portion of the plan is within the interior portion of the PTV?