Dear SDN. Question about a case I had few weeks back. Patient with PPM, pace-maker dependent undergoing sternotomy/CABG. Surgeon doesn't think "modern" pacemakers need reprogrammed or magnet so nothing was done preop. There was clear EMI/bradycardia/hypotension initially but this was transient (about 10s). It appeared that the pacemaker flipped into an asynch mode or became less sensitive to EMI and there were no issues for the rest of the case. I have found nothing in the literature to suggest that modern pacemakers are capable of this intraoperatively. My best guest is this was a bipolar lead pacemaker that already was resistant to EMI and after the chest was open (greater impedance from bovie to lead or whatever) this was no longer a problem.