Does anyone actually like Halcyon?

This forum made possible through the generous support of SDN members, donors, and sponsors. Thank you.

Anthodite

Full Member
Joined
May 17, 2024
Messages
63
Reaction score
52
Truthfully, I don’t know much about it besides that it’s the new big thing. I know it can do planning and treatment at one go and ART but that’s it. I don’t actually know any other therapist that has used it.

Now that everyone’s getting rid of their Trilogy machines, most centers I’ve seen at least have opted to get another Varian Truebeam instead, from fancy proton centers to rural areas. They always state their reason for it is the amount of patients each can take and not the price difference.

What are your opinions on using it? Is it going to be one of those machines that only work for some patients like a cyberknife/gammaknife or can it be applied to all like our current Truebeams?

Members don't see this ad.
 
Its been around for about 5 years. It can probably treat about 80% of what a typical clinic may see easily (basically anything that is IMRT/VMAT). Through put can be very fast for things like prostates as it only does 6X FFF. Many "academic" type places will have one or two these machines treating nothing but their prostates. Things that can be challenging on it are breast and anything that would be better served with electrons or require non coplanar beams or higher energy beams. It is essentially designed to be a work horse machine with a lower price point ($2 million?) compared to a True Beam.

In regards to Ethos I have never used it but know that is something that is offered and others use. I don't really understand how Ethos can work that well as the kv CBCT on Haylcon often does not produce great images.
 
Its been around for about 5 years. It can probably treat about 80% of what a typical clinic may see easily (basically anything that is IMRT/VMAT). Through put can be very fast for things like prostates as it only does 6X FFF. Many "academic" type places will have one or two these machines treating nothing but their prostates. Things that can be challenging on it are breast and anything that would be better served with electrons or require non coplanar beams or higher energy beams. It is essentially designed to be a work horse machine with a lower price point ($2 million?) compared to a True Beam.

In regards to Ethos I have never used it but know that is something that is offered and others use. I don't really understand how Ethos can work that well as the kv CBCT on Haylcon often does not produce great images.
1.5-1.8 I think

Would never own as a primary machine. Great vmat work horse though
 
Members don't see this ad :)
1.5-1.8 I think

Would never own as a primary machine. Great vmat work horse though
This.

I have no issue with the technology or the machine itself. My issues with this machine are solely practical. We have a limited number of these machines in limited locations. When they are common place, I will use them. Until then I won't.
 
This.

I have no issue with the technology or the machine itself. My issues with this machine are solely practical. We have a limited number of these machines in limited locations. When they are common place, I will use them. Until then I won't.

I think they are fairly common place at this point. Way more of these then proton machines and probably cyberknifes as well. I would guess most places with more then 3 or 4 vaults have one.
 
I think they are fairly common place at this point. Way more of these then proton machines and probably cyberknifes as well. I would guess most places with more then 3 or 4 vaults have one.
Except they only have one. Patient wants to go to other location? Re-plan. Machine goes down? Re-plan. I'm not interested in doing the same thing twice.
 
Except they only have one. Patient wants to go to other location? Re-plan. Machine goes down? Re-plan. I'm not interested in doing the same thing twice.

Those of us with non-beam matched C arm linacs at different sites that go down frequently are playing the world's smallest violin for you
 
Those of us with non-beam matched C arm linacs at different sites that go down frequently are playing the world's smallest violin for you
Replacing ours one by one with beam-matched truebeams. No room/role for halycons at a single linac site in a multi-site practice IMO
 
It can probably treat about 80% of what a typical clinic may see easily (basically anything that is IMRT/VMAT). Through put can be very fast for things like prostates as it only does 6X FFF.
Forgive me if I’m wrong, but doesn’t using the 6FFF energy mean that it’s limited to patients with a thinner body habitus and SRS/SBRTs?

From what I recall from dosimetry many prostate patients end up being an 18MV due to their weight and the 6FFF is something really utilized for SRS/SBRTs and sometimes breast radiation.
 
I think they are fairly common place at this point. Way more of these then proton machines and probably cyberknifes as well. I would guess most places with more then 3 or 4 vaults have one.
I was gonna argue but I realized I just happen live by an area concentrated with proton centers. But yeah the one place near me that got one has four vaults.
 
Its been around for about 5 years. It can probably treat about 80% of what a typical clinic may see easily (basically anything that is IMRT/VMAT). Through put can be very fast for things like prostates as it only does 6X FFF. Many "academic" type places will have one or two these machines treating nothing but their prostates. Things that can be challenging on it are breast and anything that would be better served with electrons or require non coplanar beams or higher energy beams. It is essentially designed to be a work horse machine with a lower price point ($2 million?) compared to a True Beam.

In regards to Ethos I have never used it but know that is something that is offered and others use. I don't really understand how Ethos can work that well as the kv CBCT on Haylcon often does not produce great images.

How much of the cost is the linac vs. the vault? Seems like it could be penny-wise/pound-foolish to skimp on the linac if you're already paying for the vault and staffing.
 
Members don't see this ad :)
Forgive me if I’m wrong, but doesn’t using the 6FFF energy mean that it’s limited to patients with a thinner body habitus and SRS/SBRTs?
This may be an issue indeed, if you have obese patients. We generally treat those with higher energies on C-arm linacs too.

In my experience, the Ethos in IGRT mode (which is basically a Halcyon) is excellent for pallative cases, mainly bone mets.
- The treatment couch can be lowered extensively, meaning patients with pain / mobility impairment can get on/off it very easily
- Positioning, image acquisition and treatment are very fast, meaning patients in pain only need to be on the treatment couch for a short period of time.
We use our Ethos extensively as a Halcyon exactly for these patients when we are not doing adaptive treatments.
A clinic with 3-4 linacs can easily fill up a Halcyon with palliative cases.
I am not sure if there would be a reimbursement problem in the US (since you have to treat with IMRT/VMAT on the Halcyon), for us it's not a problem.
 
How much of the cost is the linac vs. the vault? Seems like it could be penny-wise/pound-foolish to skimp on the linac if you're already paying for the vault and staffing.

Interestingly, the Halcyon requires much less shielding then a True beam as it has internal shielding and only uses a 6X energy beams.
 
Decent amount of information, decent amount of misinformation in this thread. fiji and medgator are most correct Having used both TBs and Ethos:

Truthfully, I don’t know much about it besides that it’s the new big thing. I know it can do planning and treatment at one go and ART but that’s it. I don’t actually know any other therapist that has used it.

Now that everyone’s getting rid of their Trilogy machines, most centers I’ve seen at least have opted to get another Varian Truebeam instead, from fancy proton centers to rural areas. They always state their reason for it is the amount of patients each can take and not the price difference.

What are your opinions on using it? Is it going to be one of those machines that only work for some patients like a cyberknife/gammaknife or can it be applied to all like our current Truebeams?

First off, Halcyon has been around for multiple years. Ethos is a Halcyon that has adaptive software loaded onto it. The hardware is otherwise the same, IIRC.

Trilogy is an old machine. If you are happy with Trilogy does, a TrueBeam is the upgrade that is in-line with what Halcyon does.

Halcyon/Ethos (H/E) is an excellent machine for doing most routine VMAT patients quicker in multiple ways than TB. H/E CBCT is faster than TB. Delivery is faster than TB due to 6MV-FFF beams.

The disadvantages of H/E (compared to TB) are in the following:
Most 3D treatments are worse b/c you are locked into 6MV-FFF beams.
No 6 DoF couch
Base CBCT isn't as good quality (maybe? depends on opinions) as TB, ESPECIALLY around gas - hypersight is reportedly supposed to correct this (I do not have access to it yet)
Can't do half beam blocks cleanly
No collimator rotation - just two rows of MLCs that 'act' like it
No couch kicks
Software 'requires' you do a CBCT prior to beam on every fraction, even if not reimbursable, thus can affect throughput
Max field size for a single iso about 14x14, but IMO plan quality degrades at higher ends. A longer field can be treated by doing a second iso-blended to the initial but are overall capable of giving much lower quality VMAT plans than doing a 20x20 field on a TB even with VMAT.

Forgive me if I’m wrong, but doesn’t using the 6FFF energy mean that it’s limited to patients with a thinner body habitus and SRS/SBRTs?

From what I recall from dosimetry many prostate patients end up being an 18MV due to their weight and the 6FFF is something really utilized for SRS/SBRTs and sometimes breast radiation.
Definitive prostate patients exclusively get IMRT. IMRT/VMAT with an 18MV beam I would consider borderline malpractice due to the neutron generation. Some consider IMRT/VMAT with a 10MV beam as acceptable, but many stick with 6MV (whether it is FFF or not).

6MV-FFF is primarily used for SRS/SBRT, but it can also be used with IMRT/VMAT without very significant differences in plan quality. 3D plans with 6MV-FFF look worse as the dMax is lower and thus penetrating deeper targets requires more complex beam arrangements or significantly more heterogeneity than a 10MV or 15-18MV beam.

This may be an issue indeed, if you have obese patients. We generally treat those with higher energies on C-arm linacs too.

In my experience, the Ethos in IGRT mode (which is basically a Halcyon) is excellent for pallative cases, mainly bone mets.
- The treatment couch can be lowered extensively, meaning patients with pain / mobility impairment can get on/off it very easily
- Positioning, image acquisition and treatment are very fast, meaning patients in pain only need to be on the treatment couch for a short period of time.
We use our Ethos extensively as a Halcyon exactly for these patients when we are not doing adaptive treatments.
A clinic with 3-4 linacs can easily fill up a Halcyon with palliative cases.
I am not sure if there would be a reimbursement problem in the US (since you have to treat with IMRT/VMAT on the Halcyon), for us it's not a problem.
This may be true in the non-US, but -
H/E is routinely horrible for palliative patients in the US due to frequent issues of getting IMRT/VMAT insurance approval and the fact that 6MV-FFF is trash for 3D plans for most things that require palliation, including all non-cervical spine, anything in the pelvis or abdomen. It's fine for like extremities and maybe standard whole brain.
If routinely doing IMRT for palliation, then yes, it is faster on CBCT and delivery than TB.
 
How much of the cost is the linac vs. the vault? Seems like it could be penny-wise/pound-foolish to skimp on the linac if you're already paying for the vault and staffing.

Halcyon/Ethos as a stand alone linac is a bad idea. As a second linac to speed through the IMRT/VMAT patients (a common part of most ROs practice I imagine), it's very helpful.

The idea of beam-matched linacs is nice and all and if your volumes aren't that high it's fine, but Halcyon is to get patients in and treated ASAP. Treatment times are frequently 10-15 minute time slots including the CBCT.
 
I was gonna argue but I realized I just happen live by an area concentrated with proton centers. But yeah the one place near me that got one has four vaults.
I'm surprised they didn't put a single-room Mevion Fit system instead of a Halcyon in their 4th room.

Then they too could have proton therapy.
 
3D can definitely be done on halcyon
3D Breast plans and palliative look great if have dosimetrist trained to plan on halcyon.

Agree that there is lots of misinformation on this thread.
 
Decent amount of information, decent amount of misinformation in this thread. fiji and medgator are most correct Having used both TBs and Ethos:



First off, Halcyon has been around for multiple years. Ethos is a Halcyon that has adaptive software loaded onto it. The hardware is otherwise the same, IIRC.

Trilogy is an old machine. If you are happy with Trilogy does, a TrueBeam is the upgrade that is in-line with what Halcyon does.

Halcyon/Ethos (H/E) is an excellent machine for doing most routine VMAT patients quicker in multiple ways than TB. H/E CBCT is faster than TB. Delivery is faster than TB due to 6MV-FFF beams.

The disadvantages of H/E (compared to TB) are in the following:
Most 3D treatments are worse b/c you are locked into 6MV-FFF beams.
No 6 DoF couch
Base CBCT isn't as good quality (maybe? depends on opinions) as TB, ESPECIALLY around gas - hypersight is reportedly supposed to correct this (I do not have access to it yet)
Can't do half beam blocks cleanly
No collimator rotation - just two rows of MLCs that 'act' like it
No couch kicks
Software 'requires' you do a CBCT prior to beam on every fraction, even if not reimbursable, thus can affect throughput
Max field size for a single iso about 14x14, but IMO plan quality degrades at higher ends. A longer field can be treated by doing a second iso-blended to the initial but are overall capable of giving much lower quality VMAT plans than doing a 20x20 field on a TB even with VMAT.


Definitive prostate patients exclusively get IMRT. IMRT/VMAT with an 18MV beam I would consider borderline malpractice due to the neutron generation. Some consider IMRT/VMAT with a 10MV beam as acceptable, but many stick with 6MV (whether it is FFF or not).

6MV-FFF is primarily used for SRS/SBRT, but it can also be used with IMRT/VMAT without very significant differences in plan quality. 3D plans with 6MV-FFF look worse as the dMax is lower and thus penetrating deeper targets requires more complex beam arrangements or significantly more heterogeneity than a 10MV or 15-18MV beam.


This may be true in the non-US, but -
H/E is routinely horrible for palliative patients in the US due to frequent issues of getting IMRT/VMAT insurance approval and the fact that 6MV-FFF is trash for 3D plans for most things that require palliation, including all non-cervical spine, anything in the pelvis or abdomen. It's fine for like extremities and maybe standard whole brain.
If routinely doing IMRT for palliation, then yes, it is faster on CBCT and delivery than TB.
This is a fantastic summary
I would add that as far as i know halcyon is not configured for any type of motion management. It lacks a 6dof couch

In my mind it is a workhorse for basic head and neck, whole pelvis, prostate, palliative cases.

But if i were building a new department and could put in only one machine it would be a true beam. with two, a beam matched second true beam.
 
@pikachu and @evilbooyaa have a lot of great points, but just to add since I JUST put one of these in....

-the IMRT breast plans are phenomenal, and there is cookbook JACMP paper on generating them
-6xFFF doesn't mean you can't generate good homogenous non-SBRT plans, but they will be more modulated
-big fields may need second iso & feathering, but this is not hard or time consuming on the halcyon
-the hypersight CBCT is capable of sufficient quality sims for brain, extremity, and low pelvis. stay away from lungs and anything where bowel gas impactful. there is work to do on axial boundary conditions
- no gating, but is on Varian's road map. Good breath hold coaching and gating-surrogate solution with arrays of 13field 6FFF static-field IMRT plans can still generate phenomenal SBRT plans

I thought about my dream clinic model for years, and went Halycon/Edge for a reason. In a clinic w/o much SRS, 2 beam-matched millenium true-beams might make more sense.
 
@pikachu and @evilbooyaa have a lot of great points, but just to add since I JUST put one of these in....

-the IMRT breast plans are phenomenal, and there is cookbook JACMP paper on generating them
-6xFFF doesn't mean you can't generate good homogenous non-SBRT plans, but they will be more modulated
-big fields may need second iso & feathering, but this is not hard or time consuming on the halcyon
-the hypersight CBCT is capable of sufficient quality sims for brain, extremity, and low pelvis. stay away from lungs and anything where bowel gas impactful. there is work to do on axial boundary conditions
- no gating, but is on Varian's road map. Good breath hold coaching and gating-surrogate solution with arrays of 13field 6FFF static-field IMRT plans can still generate phenomenal SBRT plans

I thought about my dream clinic model for years, and went Halycon/Edge for a reason. In a clinic w/o much SRS, 2 beam-matched millenium true-beams might make more sense.

Have you been able to get right breast IMRT approved? I usually am not.
 
Have you been able to get right breast IMRT approved? I usually am not.
You can get right breast IMRT approved w/ most insurances if you limit fraction number to 5.

Arguably, every IMRT LCD/LCA in America (the 3 LCD/LCAs cover about 20 states and D.C.) supports IMRT for any/all right breast cancers. I say this because 1) "breast" is in the thorax chapter of every med school anatomy textbook, 2) breast radiotherapy is, evidently, thoracic radiotherapy (see below), and 3) all the IMRT LCDs say IMRT is necessary for thoracic malignancies, and all the LCAs list right breast ICD-10s as supporting necessity too.

2024-11-15 11_19_02-(1) Shankar Siva on X_ _Impressive intro in French! @KathrynBanfill 🏴󠁧󠁢󠁥󠁮󠁧.png
 
Last edited:
1731805734260.png


Is this a course of radiation for a "thoracic malignancy"? I don't think we should nudge payors to cover breast IMRT as their default option. IMRT for breast Ca should be an exception.
 
I'm surprised they didn't put a single-room Mevion Fit system instead of a Halcyon in their 4th room.

Then they too could have proton therapy.
Last time I was at that center they were tryina sell investors on adaptive radiotherapy so maybe that has something to with it.
 
View attachment 395323

Is this a course of radiation for a "thoracic malignancy"? I don't think we should nudge payors to cover breast IMRT as their default option. IMRT for breast Ca should be an exception.
There is a paper in IJROBP from a couple years ago showing IMRT reduces skin toxicity vs 3D for breast CA, so there is clinical data to support its use
 
Is this a course of radiation for a "thoracic malignancy"?
No "quotes" necessary; I don't make the rules
2024-11-18 12_31_01-.png

IMRT for breast Ca should be an exception.
One timeframe's exception is another timeframe's enthusiasm.

EDIT: the most common presentation of breast cancer is stage zero/one in a 50+ yo, and per current ASTRO guidelines PBI is rec'd over whole breast for those patients (proven less side effects, proven non-inferior LC), and the most cited PBI versus whole breast trials (Livi, IMPORT-LOW, etc) all used IMRT for the PBI. So, it can be easily argued IMRT should be the standard for breast cancer. Five fraction IMRT is cheaper than 15-21 fraction non-IMRT.

2024-11-18 12_36_40-Should Intensity-Modulated Radiation Therapy Be the Standard of Care in th...png
There is a paper in IJROBP from a couple years ago showing IMRT reduces skin toxicity vs 3D for breast CA, so there is clinical data to support its use
Rad Onc really went out of its way through the years (NCCN doesn't even mention IMRT, Choosing Wisely, Ben Smith and Paul Wallner breast IMRT shaming) to shove the IMRT breast data under the rug. Reminds me kind of like when a pretty girl is totally embarrassed by her beauty, dresses homely, etc. There is, by far, more randomized data supporting IMRT for breast than randomized IMRT data in any other disease site (broken record).

https://pubmed.ncbi.nlm.nih.gov/18285602/

https://pubmed.ncbi.nlm.nih.gov/17224195/

https://pubmed.ncbi.nlm.nih.gov/33321192/

https://pubmed.ncbi.nlm.nih.gov/32980384/

https://pubmed.ncbi.nlm.nih.gov/24043742/

https://pubmed.ncbi.nlm.nih.gov/21345620/

 
Last edited:
I can only imagine the day Wallnernus came across this in his daily Netter reading and had a ‘Eureka!’ Moment. Lol.
 
Breast identifies as thoracic
OMG the day has arrived where we don’t consider breast as a separate entity.

The current ARRT guideline for radiation therapist to earn their licensure has only: “thorax:non-imrt and thorax imrt”
breast is divided up to three main types of tx(supreclav+tangents, superclav+tangents+PAB, or just tangents.)

The category breasts IMRT along with less common treatments falls under is “special” which has always confused me.

Here’s the ARRT breakdown
 
Top