Had an interesting case today. Patient rolls back into the room. Five lead goes on, shows flipped T waves. Pre-op 12 lead ECG had been normal. I get called in to take a look, for another set of eyes. Before I can comment case canceled. They grab another 12 lead when the patient is back out in holding, unchanged from pre-op, no flipped T waves.
I go back, hook myself up to the monitor (I have no know cardiac disease), change the filters, and voila! I can flip my own T waves.
[YOUTUBE]MkirL1SO5xc[/YOUTUBE]
I had known that putting a higher frequency filter on could change ST segments, but I hadn't seen it flip T waves before. The filters try to keep the baselines clean so you can diagnose arrhythmias, "Is that V-tach or a wandering baseline?". If you want to look at morphologies or segments you should turn the filters down or off.
I go back, hook myself up to the monitor (I have no know cardiac disease), change the filters, and voila! I can flip my own T waves.
[YOUTUBE]MkirL1SO5xc[/YOUTUBE]
I had known that putting a higher frequency filter on could change ST segments, but I hadn't seen it flip T waves before. The filters try to keep the baselines clean so you can diagnose arrhythmias, "Is that V-tach or a wandering baseline?". If you want to look at morphologies or segments you should turn the filters down or off.