mA to max? I start at the lowest setting and increase until I get capture, then add 10% (just as ACLS teaches with transcutaneous pacing). Only during asystole do I go with the maximum amperage, which is of course transcutaneous.
You basically want to make sure you see the capture beats and not have the pacer fire due to intrinsic beats. So set the sensitivity to asynchronous. Once you have capture, then drop the mA. Once your mA are optimized, then optimize the sensitivity. You would prefer to have the minimum number of asynchronous beats.
no references off the top of my head....i am too burnt to look it up now....i may be wrong...i am ALWAYS open to correction and learning...that is why i asked ....
Here is the Madigan Army Hospital page on floating a transvenous pacer: http://www.mamc.amedd.army.mil/cardiology/book126.pdf . I'm not sure how you would float a transvenous pacer with the mA set to the minimum. You have to achieve capture to see that it is placed properly. If your generator is set at the minimum, you have two variables determining capture, where as if you go to max, there is only one. One you have advanced the wire, achieve capture, and are sitting at the RV apex, then I'd turn it down to the minimum and then increase by 50-100%.