What is the appropriate order to note vent settings in the ICU?
I have been told..
Mode/FiO2/RATE/PEEP
is incorrect

Members don't see this ad.
 
What is the appropriate order to note vent settings in the ICU?
I have been told..
Mode/FiO2/RATE/PEEP
is incorrect

I'd start with mode, but what you do next often depends on the mode.

After mode, go with rate or volume, depending on the vent setting.
...pressure control, I'd lead with volume then peak pressure then rate
...rate/volume control, I'd lead with rate then TV
...FiO2 next
...then PEEP

To be quite honest, though, I'd ask the person who told you what you were doing was wrong. Clearly they have certain demands.
 
It depends on the facility....the two most common I've seen:
For volume ventilation-
Mode/FiO2/TV/Rate/PEEP

For pressure ventilation-
Mode/FiO2/Peak Pressure/Rate/PEEP

You're best bet is to ask the person that told you were making an error or to ask the RTs on staff at the facility as they would know what is expected.
 
Members don't see this ad :)
It depends on the facility....the two most

For pressure ventilation-
Mode/FiO2/Peak Pressure/Rate/PEEP

This is common, but nowadays, we don't set the oxygen in the NICU but have standing weaning orders.

Sometimes we'll write the pressure as peak/peep as in 25/5, in which case we would write

"sIMV, 25/5, rate - 40 and 50% Fi02, wean 02 per protocol" type of orders. If on pressure support, this is added after the pressure orders as in "25/5, PS -12, IMV - 30, Fi02 -50%"

and DKM - we need you on the neo forum, so play nice in your posts on pre-allo so you don't get booted again....
 
and DKM - we need you on the neo forum, so play nice in your posts on pre-allo so you don't get booted again....

It's nice to know I'm needed. 😀 BTW check the Lounge....I'm trying to learn to be more humble and would appreciate advice.
 
ICU's I've worked at (ex ICU nurse) do it like this ...

Mode, Rate, Tidal Volume ---- Then FiO2, peep
SIMV 14 X 550ml ..... 50% O2 and 10cmH2O peep, you could then talk about pressures generated, and ABGs.the first bit effects the patient's CO2, the second bit effects the O2


if it's a PC mode, it's

Mode, Rate, PC level above peep.... then .... Fio2 and Peep.... you could then talk about volumes generated, and ABGs

i cant wait to get back to ICU
 
ICU's I've worked at (ex ICU nurse) do it like this ...

Mode, Rate, Tidal Volume ---- Then FiO2, peep
SIMV 14 X 550ml ..... 50% O2 and 10cmH2O peep, you could then talk about pressures generated, and ABGs.the first bit effects the patient's CO2, the second bit effects the O2


if it's a PC mode, it's

Mode, Rate, PC level above peep.... then .... Fio2 and Peep.... you could then talk about volumes generated, and ABGs

i cant wait to get back to ICU
Thats how I did it. I wrote my orders to mimic my notes. Maybe it's a regional thing? The attending I mostly worked at come from Colorado.
 