Here's what I would do.
First of all, I really don't like the idea of going digging under the chest of a prone patient to put magnets on or take them off in the heat of a critical event, nor do I like the idea of the pressure of the magnet over the chest for the whole case. I want the rep to disable the antitachy therapy preop. This is an ICD placed for primary prevention, and if there is no history of arrhythmia, it is unlikely to happen during the surgery. However "unlikely" is not the same as "impossible," so we place the pads. IF there was a strong history of being shocked, I would consider doing the magnet thing so that I could remove it to let the device shock him internally if it came to it. I think it'd be important to make sure that removing the magnet actually re-enables the antitachy therapy. Have the rep confirm this.
With respect to the pacemaker. We have a dual chamber device, and we see pacing spikes with a rate of 70. It is entirely possible that this is a sinus rhythm at 70, with atrial sensing and ventricular pacing. We should find out. Is there one pacing spike or two? If only one spike, is there a p-wave before every spike?
It is also entirely possible that his native rhythms are slow or nonexistant, so that the device is set to a backup rate of 70, making it essentially a DOO device. He may even be in afib, with a device essentially functioning in VVI mode at a rate of 70.
Why does it matter? I want to know because if he's atrially tracking, then he goes about his life with his HR going up and down like the rest of us. And if he was to experience blood loss, his HR would normally go up to compensate.
If you just asynch him DOO at 70, you'll probably make it through the case OK, but I wouldn't say you were optimizing his hemodynamics, since you're marrying him to the rate he's in at rest. If I know I'm working with a sadistic butcher, I may have the rep bump the rate up to 80 or 85 preop, just to give me a little extra room on the cardiac output I have to work with if we do lose some volume. I'd probably figure out (by interrogating the device) what kind of rate he produces when he exerts himself, and aim for something in that range for the case.
This approach is probably less science and more art, but I think it's a more physiologically appropriate approach to the procedure.