Anesthesia machine post COVID positive patient.

This forum made possible through the generous support of SDN members, donors, and sponsors. Thank you.

Clinicall

Full Member
10+ Year Member
Joined
Jan 9, 2013
Messages
257
Reaction score
138
Wondering what the transmission possibility would be, if any, of an anesthesia machine infecting another patient following a patient who had COVID. Not sure if this has been discussed before.

Members don't see this ad.
 

Short Answer:

  • Place a “high quality” viral filter between the breathing circuit and the patient’s airway with the capability to sample gas from the machine side of the filter.
    • HMEF is preferred to preserve humidification.
    • If filter only is used, reducing fresh gas flow is an important strategy for preserving humidity. (1-2 L/min or less)
  • Place a second filter at the end of the expiratory limb at the connection to the anesthesia machine.
  • Local conditions of filter availability will determine what devices can be employed. Alternatives are described below.
  • Breathing circuits should be discarded after every patient.
(See next FAQ for details on which filtration devices to use and special considerations for small patients)

There are two reasons to protect the anesthesia machine from contamination by a potentially infected patient. First, if pathogens are allowed to enter the internal parts of the machine, they could be passed on to a subsequent patient. Second, respiratory gases sampled for gas analysis can pass pathogens on to other patients or healthcare workers after leaving the gas analyzer if not managed properly.

The good news is that the same precautions can be applied to all patients. The strategy is not different depending upon the patient’s risk of infection. A “high quality” filter placed between the breathing circuit and the patient’s airway will protect the machine from contamination and also gases sampled for analysis. Heat and moisture exchange filters (HMEFs) are a good choice since they preserve airway humidity, and are designed so that sampled gas is filtered before it enters the gas analyzer (note that HMEs without filters provide no protection to the anesthesia machine or gas analyzer). It is possible to use a filter only at the airway that is not also a heat and moisture exchanger, but strategies like low flow anesthesia should be employed to preserve humidity. If a filter only is used at the airway, lower fresh gas flows (1-2 L/min or less) are desirable during maintenance of anesthesia to preserve humidity in the circuit.

Preferred Viral Filter Configuration

While a high quality filter at the airway will protect both the sampled gas and the anesthesia machine, there are reasons to consider adding a second filter at the end of the expiratory limb. The airway filter can become soiled and lose effectiveness and in small patients, airway filters are designed to minimize dead space and as a result are generally less effective. Not only is the second filter a reasonable backup to protect the machine from any particles that pass the primary filter, but it significantly amplifies the effectiveness of the first filter.
 
Members don't see this ad :)
We're using a circuit with viral filters on both sides of the circuit as well as between the circuit itself and airway (3 total filters). Initially we had to use our supply of viral filters plus the circuit, but now we buy the whole thing pre-packaged from the manufacturer.
 
  • Like
Reactions: 2 users
Wondering what the transmission possibility would be, if any, of an anesthesia machine infecting another patient following a patient who had COVID. Not sure if this has been discussed before.

1, we usually put COVID patients as last case of that OR
2, we try to avoid GA whenever possible due to aerosolization risk
3, we place 2x viral filters on, i'm told it reduces the cross contamination risk to zero
 