I'm getting confused about zeroing and transducer placement for arterial line. From my understanding, at least where I'm at, they zero using stopcock close to where the transducer is. Once it is zeroed, we don't need to re-zero but just move the transducer according to pt. Is this true?
The following is the question from Hall's book.
20-gauge arterial line connected to a transducer that is located 20 cm below the level of the heart. The system is zeroed at the stopcock located at the wrist while the patient’s arm is stretched out on an arm board. How will the arterial line pressure compare with the true blood pressure (BP)?
A. It will be 20 mm Hg higher
B. It will be 15 mm Hg higher
C. It will be the same
D. It will be 15 mm Hg lower
I thought it would be B but correct answer is C. Is the BP the same because it was zeroed and transducer/pt's position never changed thus BP reading is accurate.
If so, if I were to move the pt's bed higher, in order to get accurate reading, I either have to rezero and not move the transducer OR move up the transducer the same distance as pt's bed?
Thank you
The following is the question from Hall's book.
20-gauge arterial line connected to a transducer that is located 20 cm below the level of the heart. The system is zeroed at the stopcock located at the wrist while the patient’s arm is stretched out on an arm board. How will the arterial line pressure compare with the true blood pressure (BP)?
A. It will be 20 mm Hg higher
B. It will be 15 mm Hg higher
C. It will be the same
D. It will be 15 mm Hg lower
I thought it would be B but correct answer is C. Is the BP the same because it was zeroed and transducer/pt's position never changed thus BP reading is accurate.
If so, if I were to move the pt's bed higher, in order to get accurate reading, I either have to rezero and not move the transducer OR move up the transducer the same distance as pt's bed?
Thank you