Sure, but there are still recommendations to interrogate post magnet at the least.
Ideally you'd have the rep interrogate pre-procedure, deactivate tachy therapy and set to whatever asynchronous rate you want if pacer dependent and likely to be interfered with by EMI, and then return to prior settings postop.
The practice of simply using a magnet on every CIED because manufacturer spec sheets tell you the expected action and that there will be flawless return to baseline with magnet removal is potentially unsafe. It is possible for the magnet response to have been deactivated, the device to not respond as you expect, or failure to revert to pre-procedure settings with magnet removal (magnet/EMI can damage or reset existing settings, the response to magnet and/or magnet removal can be affected by battery level, etc).
It's simply a patient safety issue. Not to mention how it would look if you magnet deactivated for a surgery then removed the magnet and discharged the patient without interrogating the device and the patient had a bad outcome at home due to a change to or outright failure of device intervention/therapy.