Just curious if any of you had thoughts on using IMRT in the lungs. There does not seem to be much evidence either way. Do any academic centers use IMRT consistently for lung cancer?
Are you you looking at lung V5 or any other parameters beyond MLD and V20 with IMRT?
Yeah V5<50, V10<40, if possible. I think there is some Chinese data on V10<50 predicting for lower rates of pneumonitis - http://www.ncbi.nlm.nih.gov/pubmed/20462424?dopt=Abstract.
Been trying to keep mean lung lower than 15 if possible, as well.
IMRT is less likely to augment efficacy (tumor control) for Stage III NSCLC if dose escalation does not improve outcomes ... IMRT can improve lung dose-volume measures and can allow delivery of 60+ Gy for tumors near the spine, for which there are decades worth of data demonstrating the importance (i.e. lower V20, maintain safe cord dose). For low dose lung exposure, IMRT leads to an increase, but this tends to be more so for V<5.Thank you all for your responses -- it is really interesting to me that such a fundamental question in our field has not been tested more rigorously -- especially given all the differences (dose rate, interplay, incidental nodal radiation, conformality at the edges) that could have a bearing on local control.
I think the worry is not whether imrt augments disease control but actually whether it is inferior to 3dcrt. .
And this is precisely why there will never be a randomized trial comparing the two. It is a tremendous financial disincentive, just like for protons.
Personally, I think IMRT for lung cancer should be reserved in cases where lung, spinal cord, or brachial plexus constraints cannot be met with 3DCRT.
I think part of the challenge of designing such a trial is the question of what the most optimal plan is; often times I can get a 3D-CRT plan to meet all the constraints I am looking for but when I compare to an IMRT plan it not only meets all the constraints but improves target coverage while reducing esophageal dose, V20, and MLD (usually these are N2 cases). Its hard to assess for plan quality in a trial when you are simply looking ot meet constraints and not maximize out the efficiency of IMRT.
In terms of the value of IMRT, I don't think its likely that IMRT would impact clinical outcomes compared with 3D-CRT but I do imagine a benefit in terms of pneumonitis, esophagitis, etc. Problem is that in my experience, different institutions code these things very differently and it would be almost impossible to get uniform reporting of toxicities over a 1-2 years which would be needed for the pneumonitis info.
The type of IMRT used plays a majoe role in dose uncertainties with moving targets (such as lung cancer). Interplay effect is the key term here.
Rapid Arc / VMAT is considerably more prone to dose uncertainties than step-and-shoot IMRT. Of course, fractionation also plays a role.
Well... that's IMO debatable. When claiming that IMRT is less prone to interplay effect than VMAT, have to specify what type of IMRT you're talking about - sliding window or step-and-shoot?
The type of IMRT used plays a majoe role in dose uncertainties with moving targets (such as lung cancer). Interplay effect is the key term here.
Rapid Arc / VMAT is considerably more prone to dose uncertainties than step-and-shoot IMRT. Of course, fractionation also plays a role.
There may be more dose uncertainties with these, but that shouldn't dismiss the uncertainty of step and shoot IMRT with a tumor that may show a lot of movement with breathing.
Ok, here's some literature on the subject:
Court et al., Med. Phys. 35 (2008) 1926-1931
In this graph from the paper you can see phantom measurements demonstrating, that RapidArc results in the largest dose uncertainties compared with step-and-shoot IMRT.
![]()
This is interesting data, however, I somewhat disagree with your interpretation. I think this suggests that once you reach about 20 fractions (which would be a fairly low amount in a definitive lung cancer case), RapidArc (single or double), dynamic IMRT (simple, complex, hybrid, 200-600), step and shoot IMRT, conformal, and pretty much everything else are all the same (not inferior), with the sole exception of a complex single rapid arc.