3D... been burned by IMRT a few times.
 
Just did one that finished last week. 5 field 3D CRT. Not sure you could get the insurance to cover imrt as none of the OAR were anywhere close to dose limits. Had the patient sip a few mL's of contrast prior to treatment with kv CBCT before each fraction to verify target position/coverage.
 
  • Like
Reactions: 1 users
Members don't see this ad :)
3D.

As a VMAT-era RadOnc, there's not a lot of definitive cases I use 3D for. Indolent lymphoma is my primary indication. Usually in the form of follicular or gastric MALT.

ISRT for Hodgkin's I use IMRT/VMAT usually because of how young they are. ISRT for DLBCL depends on the case.
 
  • Like
Reactions: 1 user
More important than technique is the image guidance imo.. Daily CBCT, fast 3-4 hours before sim and daily treatment. Can get a decent plan either way. Medicare will pay for IMRT if you think it looks better
 
  • Like
Reactions: 11 users
I've done both. Agree with Medgator.
 
More important than technique is the image guidance imo.. Daily CBCT, fast 3-4 hours before sim and daily treatment. Can get a decent plan either way. Medicare will pay for IMRT if you think it looks better

Bingo. My practice as well.
 
  • Like
Reactions: 1 users
3D... been burned by IMRT a few times.
Which are people generally doing?
I would do a 4 field static beam IMRT plan and turn the smoothing up on each beam. I could just make a nicer bit more homogenous plan than what you guys are calling 3D. I would also call this IMRT approach 3D. :) And I can make the plan so darn quickly. And if you wanna tweak the kidney doses or whatever… flexible and efficient imho. Stomach has a lot of day to day variation. I will sim on separate days to make an ITV from fused sims. Daily CBCT is comforting.
 
Explain ? Curious to understand how I could screw this up
I had one patient with MALT who had a distended stomach at baseline (I think some component of gastroperesis 2/2 diabetes) and, despite being NPO, his stomach volume was vacillate dramatically from day to day. I had originally planned him VMAT because his stomach was so large, we were starting to get into kidney territory... but I ended up switching him to 3D because the highly conformal VMAT plan was too tight some days on CBCT.
 
  • Like
Reactions: 1 user
Members don't see this ad :)
I had one patient with MALT who had a distended stomach at baseline (I think some component of gastroperesis 2/2 diabetes) and, despite being NPO, his stomach volume was vacillate dramatically from day to day. I had originally planned him VMAT because his stomach was so large, we were starting to get into kidney territory... but I ended up switching him to 3D because the highly conformal VMAT plan was too tight some days on CBCT.
Doesn’t sound like a technique issue?
Motion is a problem always and 3D doesn’t fix that - you’re just using bigger margins, right?

This at the end of the day is about target delineation not technique. But, I think people make it about technique. IMRT isn’t failing the patient. Inadequate motion management/stomach distension exist with 3D, too.
 
  • Love
Reactions: 1 user
Doesn’t sound like a technique issue?
Motion is a problem always and 3D doesn’t fix that - you’re just using bigger margins, right?
I mean... I could have thrown on a 2 cm margin and still done VMAT, but 3D worked just fine... just had to get creative with field arrangement. Sometimes a shotgun is better than a riffle (although, I don't have much experience with either haha)
 
  • Like
Reactions: 1 user
I mean... I could have thrown on a 2 cm margin and still done VMAT, but 3D worked just fine... just had to get creative with field arrangement. Sometimes a shotgun is better than a riffle (although, I don't have much experience with either haha)
That makes sense. I don’t know the gun analogy well!
 
If you ask ASTRO - there's only one answer.

PROTON.
 
  • Like
Reactions: 1 user
  • Like
Reactions: 1 users
I had one patient with MALT who had a distended stomach at baseline (I think some component of gastroperesis 2/2 diabetes) and, despite being NPO, his stomach volume was vacillate dramatically from day to day. I had originally planned him VMAT because his stomach was so large, we were starting to get into kidney territory... but I ended up switching him to 3D because the highly conformal VMAT plan was too tight some days on CBCT.
This wasn't a failure of modality but of margins (as you I'm sure know!); a preparation error and not an execution error.

EDIT: didn't see what Simul wrote above
 
Agree w/ depends on anatomy, can do 3D or VMAT. 3D might be better if can stay off kidneys.
You kill me with this @evilbooyaa ! You’re a young guy - how is it better ? Our faculty did a real number on us about the value of treatment technique.

If the target is the same - which it should be - I’m curious as to what parameter would be better with 3D? If you did 4-5 static imrt fields, maybe V5 marginally better.

I think I just signaled Ryckman
 
  • Like
  • Haha
Reactions: 1 users
You kill me with this @evilbooyaa ! You’re a young guy - how is it better ? Our faculty did a real number on us about the value of treatment technique.

If the target is the same - which it should be - I’m curious as to what parameter would be better with 3D? If you did 4-5 static imrt fields, maybe V5 marginally better.

I think I just signaled Ryckman

3D might be better than VMAT is what I said. As I'm sure you're aware, static IMRT != VMAT. VMAT will likely lead to higher kidney doses than 3D. I do not routinely using static IMRT in 2022 because if I'm doing 3D I'm usually getting a good homogeneous plan, which is the main thing static IMRT can improve upon (same concept as sliding window) that I am comfortable with.

If I feel like I can get decent homogeneity with a 3 or 4-field box w/ 3D why would I bother with static IMRT? Especially for 30Gy? When the main reason patients get symptoms is due to radiation of the target itself (the stomach)? 3D can have a treatment that takes a shorter amount of time to deliver with an otherwise functionally equivalent plan.
 
  • Like
Reactions: 1 users
You kill me with this @evilbooyaa ! You’re a young guy - how is it better ? Our faculty did a real number on us about the value of treatment technique.

If the target is the same - which it should be - I’m curious as to what parameter would be better with 3D? If you did 4-5 static imrt fields, maybe V5 marginally better.

I think I just signaled Ryckman
IMO, you can do either and the potential to reduce “near misses” with 3D depends on how you do it. The stomach has a lot more degrees of freedom to move/rotate and distort than a lot of organs. If you use a 4-field box, you will have more wiggle room in at least some directions than with IMRT. Are they the right directions? Not a guarantee but there will be more wiggle room. Now if you use more fields and get super conformal…not so much. Might as well do VMAT and make it a fast delivery.

I’ve done both with good results. Personally, I find the set up more reliable if you have them fast and then drink a small volume of water before sim and treatment (100-150 cc). Tummies don’t drain on a schedule and an empty stomach is still somewhat variable.
 
  • Like
Reactions: 1 users
Tummies don’t drain on a schedule
They do in my clinic:

1652139365703.png
 
That’s reasonable. But I don’t think I’ve ever seen better dosimetry in the abdomen.
3D might be better than VMAT is what I said. As I'm sure you're aware, static IMRT != VMAT. VMAT will likely lead to higher kidney doses than 3D. I do not routinely using static IMRT in 2022 because if I'm doing 3D I'm usually getting a good homogeneous plan, which is the main thing static IMRT can improve upon (same concept as sliding window) that I am comfortable with.

If I feel like I can get decent homogeneity with a 3 or 4-field box w/ 3D why would I bother with static IMRT? Especially for 30Gy? When the main reason patients get symptoms is due to radiation of the target itself (the stomach)? 3D can have a treatment that takes a shorter amount of time to deliver with an otherwise functionally equivalent plan.
VMAT is a type of IMRT
Like saying radiation does not equal protons.
At the risk of going full Scarb, I get it. I’m a stickler for the fact that basically it’s all just manipulation of fluence and the target is the target is the target.

ESE/Scarb - dose cloud , probabilities , blah blah blah
 
  • Like
Reactions: 1 user
That’s reasonable. But I don’t think I’ve ever seen better dosimetry in the abdomen.

VMAT is a type of IMRT
Like saying radiation does not equal protons.
At the risk of going full Scarb, I get it. I’m a stickler for the fact that basically it’s all just manipulation of fluence and the target is the target is the target.

ESE/Scarb - dose cloud , probabilities , blah blah blah
Also, static IMRT is not the same as sliding window and the goal of it is not generally yo improve homogeneity. How do you think we did IMRT for everything way back when? 5,7,9 static fields to create concave shaped fields.
 
That’s reasonable. But I don’t think I’ve ever seen better dosimetry in the abdomen.

VMAT is a type of IMRT
Like saying radiation does not equal protons.
At the risk of going full Scarb, I get it. I’m a stickler for the fact that basically it’s all just manipulation of fluence and the target is the target is the target.

ESE/Scarb - dose cloud , probabilities , blah blah blah
Is field-in-field IMRT? Lol sorry, couldn’t miss a chance to troll
 
  • Like
  • Haha
Reactions: 4 users
That’s reasonable. But I don’t think I’ve ever seen better dosimetry in the abdomen.

VMAT is a type of IMRT
Like saying radiation does not equal protons.
At the risk of going full Scarb, I get it. I’m a stickler for the fact that basically it’s all just manipulation of fluence and the target is the target is the target.

ESE/Scarb - dose cloud , probabilities , blah blah blah
1652148564358.png
 
  • Like
  • Haha
Reactions: 1 users
As I'm sure you're aware, static IMRT != VMAT.
Huh???

I'm sure we all know what we mean "inside our head," but we have to speak the same lingo.

Travel with me back to 2002 or 2003. When I say dynamic MLC, aka dMLC, this is sliding window. Sliding window was a bit of an improvement over step-and-shoot IMRT. Sliding window (usually contrasted with "step and shoot" or segmental IMRT) allowed more "segments" essentially, which allowed more inhomogeneity and resolution *within* the beam's fluence. Beams were always static beams ie non-arc'ed. No one really said "static IMRT" back then. Step-and-shoot is not: do a beam, then move to another gantry angle and do a different beam, etc. I mean it "looks like" you're stepping and shooting, which you are, but step and shoot refers to the MLCs stepping, then shooting, versus constantly in motion and gliding, aka dMLC.

UheSNSV.png


I0L7HAn.png

If you use a 4-field box, you will have more wiggle room in at least some directions than with IMRT.

You can make a 4-field IMRT box. So in some sense your sentence reads:

If you use a 4-field box, you will have more wiggle room in at least some directions than with a 4-field box.
 
Last edited:
  • Love
Reactions: 1 user
Yes, IMRT is an over-arching thing that encompasses lots of things. All VMAT is IMRT, but not all IMRT is not VMAT. Thus, when I said VMAT, I am talking about a small proportion of IMRT.

Not going to get in an argument about semantics, because I'm not perfect on the semantics, as I still at times use the word hotspot (instead of heterogeneity) when it relates to SRS or SBRT plans.

Static field IMRT is, to me, when you pick your beam angles and each of them does IMRT. So I suppose 'Fixed Gantry' IMRT rather than 'Dnamic Gantry'. So fine, re-phrase my statement to 'Fixed Gantry' IMRT.

I will withdraw my statement of static IMRT (or Fixed Gantry IMRT) doing the same thing as sliding window, as I realize that you could modulate to carve dose more. However, my point for the initial case stands. VMAT will (usually) spray more low dose around where you are trying to treat. For a gastric MALT, there is frequently no OAR immediately adjacent to the target that requires significant blocking, especially at the Rx dose.

Whether it is done as step and shoot (I learned from SDN that that refers to the MLC movement, not the gantry movement and have since re-framed my terminology as that seems to be a common misconception on semantics as well) or dynamic MLC won't make a clinical difference, but patient might prefer dynamic MLC for what (I presume to be) a slightly faster treatment time.
 
  • Like
  • Love
Reactions: 2 users
I've learned more about tx planning and dosimetry on SDN than I did in residency.
 
  • Like
  • Love
Reactions: 5 users
I've learned more about tx planning and dosimetry on SDN than I did in residency.
Same as a resident, and I can only hope to pass that along to future generations, with the goal being to educate without as much of the raised eye-brows semantics discussion (understanding that meanings of words change over time to fall into line with local colloquialisms, AKA hotspot) that plagues a few of the regulars. There were a few of the original zombies, others have been more recently infected.
 
  • Like
Reactions: 1 user
There should be a 4 field box option in arc planning as this is totally achievable and probably quicker to plan and prettier, and qa these days using portal dosimetry is probably quicker than shooting 4 ports. Then everyone can be happy
 
There should be a 4 field box option in arc planning as this is totally achievable and probably quicker to plan and prettier, and qa these days using portal dosimetry is probably quicker than shooting 4 ports. Then everyone can be happy
I feel like you are mocking me :)
 
Thinking hot spot is a problem to say is part of the problem! That’s a made up semantic rule

It is a hot spot! The difference is with SBRT we want hot spots and with standard VMAT people don’t like them (and they should be fine with hot spots, that’s a boomer vestige to not want hot spots in a chemoRT plan IMO) but it absolutely is by definition a hot spot.
 
Kinda joking. Can open the leaves from 358 to 2 degrees, 88-92 degrees, 178-182 degrees, and 268-272 degrees. Would be a nice "4 field box."

Like this version of AP/PA
 
  • Like
Reactions: 1 user
Kinda joking. Can open the leaves from 358 to 2 degrees, 88-92 degrees, 178-182 degrees, and 268-272 degrees. Would be a nice "4 field box."

Like this version of AP/PA
there is something about the hardware engineering that the gantry would have to move so slowly to "spray" the fluences over just a 178-182 degree travel (also, isn't this an impossible arc, because the gantry can't slide past about 181 on the downward side, at least on a Varian machine... ethos and halcyon may be different idk if they are "slip gantries") that something doesn't work mechanically. I think the thing I'm thinking of is... MUs per degree??? Too many MUs per degree with short-angle arcs? Anyways, the minimum arc is usually 30 degrees from start angle to stop angle: to wit, your link...
GeIK7kh.png
 
Thinking hot spot is a problem to say is part of the problem! That’s a made up semantic rule

It is a hot spot! The difference is with SBRT we want hot spots and with standard VMAT people don’t like them (and they should be fine with hot spots, that’s a boomer vestige to not want hot spots in a chemoRT plan IMO) but it absolutely is by definition a hot spot.
By definition it is not a hot spot.
jHKuY8N.png


When you say "hot spot," what if in my head I think a hot region outside the target? Because that is the definition of hot spot. What you are talking about is inhomogeneity. I mean, worst case, you tell a dosimetrist "I like a plan with a little hot spot," and the dosimetrist follows your instruction per the actual definition of hot spot ;)

Nah, I do think semantics matter. "They should be fine with hot spots..." uh, like, NEVER accept a hot spot in a single iso small met SRS cranial uncomplicated case. It's like in a sci-fi movie when someone says "That event was more than 100 light-years ago." Then I can't even watch the movie anymore.
 
Fair enough, if you’re a semantics person I think that has value.

I would add that I think many people use hot spot to refer to hot spots in the PTV is good for even VMAT, learned today that is ‘wrong’, good to know, in case I ever find myself in AAPM jail.
 
  • Like
Reactions: 1 user
there is something about the hardware engineering that the gantry would have to move so slowly to "spray" the fluences over just a 178-182 degree travel (also, isn't this an impossible arc, because the gantry can't slide past about 181 on the downward side, at least on a Varian machine... ethos and halcyon may be different idk if they are "slip gantries") that something doesn't work mechanically. I think the thing I'm thinking of is... MUs per degree??? Too many MUs per degree with short-angle arcs? Anyways, the minimum arc is usually 30 degrees from start angle to stop angle: to wit, your link...
GeIK7kh.png
I can never remember where they start, but if so, 180-78 and then 182-180 or therabout. I didn't say we yet possess the tech to entirely recreate a 4 field box in a single arc, but this doesn't sound hard. It's worth finding out.
 
  • Like
Reactions: 1 user