Sorry I may have missed it can someone point me to the thread with the 25G technique? Thanks.
Do you just use the blank wire to tell when you're in?I've done several dozen SCS trials and implants with no LOR.
Posterior and midline in the CLO view is consistent. The angle of the lead relative to the lamina. I get a lateral early. AP to steer.Do you just use the blank wire to tell when you're in?
I hate the false lead LOR as it goes up 2-3 contacts and curls backwards behind the flavum.Posterior and midline in the CLO view is consistent. The angle of the lead relative to the lamina. I get a lateral early. AP to steer.
Done this many times.
Same technique as the 25g CESI, but the lead is the contrast. Needle tip 1-2mm posterior to LF, stylet out, lead in...Microadvancements of the needle with slight pressure on the lead.
Very crisp feedback as you poke through with your lead.
Careful that your lead isn't sticking so far out of the needle that the LF denies the lead epidural entry and the lead folds over on itself.
You should have around 1-2 electrodes or so distance out of the needle so there is enough strength in the lead not to bend too much.
You COULD fracture your lead that way, especially if you're clumsy and ham fisting your lead against the lamina itself.
What angle oblique do you find optimal at T12-L1 for CLO?Posterior and midline in the CLO view is consistent. The angle of the lead relative to the lamina. I get a lateral early. AP to steer.
Done this many times.
Same technique as the 25g CESI, but the lead is the contrast. Needle tip 1-2mm posterior to LF, stylet out, lead in...Microadvancements of the needle with slight pressure on the lead.
Very crisp feedback as you poke through with your lead.
Careful that your lead isn't sticking so far out of the needle that the LF denies the lead epidural entry and the lead folds over on itself.
You should have around 1-2 electrodes or so distance out of the needle so there is enough strength in the lead not to bend too much.
You COULD fracture your lead that way, especially if you're clumsy and ham fisting your lead against the lamina itself.
I go full lateral.What angle oblique do you find optimal at T12-L1 for CLO?
It's 45-50ish for me.What angle oblique do you find optimal at T12-L1 for CLO?
Damn really?? Cheaper too. Saline or air?Once I went 3ml plastic syringe as LOR I never went back.
Glass LoR is personally offensive to me.
Kids today.when i see a fellow or a young attending break out the glass LOR syringe, then bathe it in saline and pump it a few times..... a little piece of me dies inside