When you give 6% des with 2 L of O2, what is the FiO2?

100 or 94%
 

jwk

CAA, ASA-PAC Contributor
15+ Year Member
Apr 30, 2004
3,640
764
Atlanta, GA
When you give 6% des with 2 L of O2, what is the FiO2?

100 or 94%
In a perfect world, it should be 94% at equilibrium.

If you were giving 2L O2 and 2L N20, it would be 50% (yes, I know you know).

The inspired fraction (Fi) by definition can't be more than 1 or 100%.
 
OP
urge

urge

10+ Year Member
Jun 23, 2007
3,784
1,160
Status
Attending Physician
Why is it everybody charts 6% des with a FiO2 of 1.0?
 

Pilot Doc

SDN Angel
Moderator Emeritus
15+ Year Member
Mar 6, 2002
1,649
7
Visit site
Status
Resident [Any Field]
convenience.

The difference between 100% and 94% in partial pressure of oxygen is within the range of difference you could see with atmospheric pressure variation. the difference isn't significant.
 

pgg

Laugh at me, will they?
Administrator
10+ Year Member
Dec 15, 2005
12,112
8,045
Home Again
Status
Attending Physician
urge said:
When you give 6% des with 2 L of O2, what is the FiO2?

100 or 94%
urge said:
Why is it everybody charts 6% des with a FiO2 of 1.0?
Most people I know chart what the O2 sensor or gas analyzer says, which in this case would probably be around 90% (H2O vapor, left over nitrogen accounting for the rest).

But yeah, it bugs me too when I take over a case and there's a string of 1s.
 

Planktonmd

Moderator Emeritus
Lifetime Donor
10+ Year Member
Nov 2, 2006
6,973
2,077
The South
Status
Attending Physician
Only the newest generation of vaporizers are altitude compensated. (Except desflurane)
True, but as long as you don't change altitude during the administration of anesthesia (you don't give anesthesia on an unpressurized airplane I hope), if you maintain a constant partial pressure you should expect a constant percentage of the volume.
 

periopdoc

Cardiac Anesthesiologist
Lifetime Donor
10+ Year Member
Sep 8, 2008
2,253
321
Kalispell, Montana
Status
Attending Physician
Only the newest generation of vaporizers are altitude compensated. (Except desflurane)
What do you mean by "altitude compensated." Aren't all modern variable bypass vaporizers inherently altitude compensated by the fact that turning the dial adjusts the partial pressure output? (The dial says percent, but you are actually adjusting the partial pressure output of the vaporizer). You will deliver a higher volume percent at altitude, but you need to to maintain the desired pressure. Because of this, a setting of 1 on an isoflurane vaporizer at sea level has an identical effect as a setting of 1 at altitude. Given identical patients etc)

As you note, Desflurane is not administered from a variable bypass vaporizer. Instead the dial sets a volume percent so you need to set a higher number at altitude for the same effect. A setting of 7 at altitude will be less effective than at sea level.

For the mathematically inclined, here is the vapor output equation.

VO= (CGxSVP)/(Pb-SVP)

VO=vapor output (ml), CG= carrier gas flow(mL.min), SVP=saturated vapor pressure (mm Hg) at room temp, and Pb is barometric pressure (mm Hg)




- pod
 

jwk

CAA, ASA-PAC Contributor
15+ Year Member
Apr 30, 2004
3,640
764
Atlanta, GA
Most people I know chart what the O2 sensor or gas analyzer says, which in this case would probably be around 90% (H2O vapor, left over nitrogen accounting for the rest).

But yeah, it bugs me too when I take over a case and there's a string of 1s.
And that's the problem.

FiO2 is a measurement for us, just like SaO2, EtCO2, temp, etc. Every anesthesia setup is required to have at least one, if not more, FiO2 monitors. The number that should be going on your anesthesia record is the number on the monitor. If you're writing down 1, or 100%, or just estimating based on your O2 and N2O concentrations only, you might as well not be writing it down at all because it's simply wrong.
 

Pilot Doc

SDN Angel
Moderator Emeritus
15+ Year Member
Mar 6, 2002
1,649
7
Visit site
Status
Resident [Any Field]
If you're writing down 1, or 100%, or just estimating based on your O2 and N2O concentrations only, you might as well not be writing it down at all because it's simply wrong.
I suspect most people watch the number bounce around between 94 and 97 and just write down 1.0 once an hour. It might be imprecise, but it is hardly wrong.

Atmospheric pressure varies over a 10% range. It's not documented at all.
That's more imprecision than the 100% rounding.
 

pgg

Laugh at me, will they?
Administrator
10+ Year Member
Dec 15, 2005
12,112
8,045
Home Again
Status
Attending Physician
And that's the problem.

FiO2 is a measurement for us, just like SaO2, EtCO2, temp, etc. Every anesthesia setup is required to have at least one, if not more, FiO2 monitors. The number that should be going on your anesthesia record is the number on the monitor. If you're writing down 1, or 100%, or just estimating based on your O2 and N2O concentrations only, you might as well not be writing it down at all because it's simply wrong.
Well put.
 

pgg

Laugh at me, will they?
Administrator
10+ Year Member
Dec 15, 2005
12,112
8,045
Home Again
Status
Attending Physician
I suspect most people watch the number bounce around between 94 and 97 and just write down 1.0 once an hour. It might be imprecise, but it is hardly wrong.

Atmospheric pressure varies over a 10% range. It's not documented at all.
That's more imprecision than the 100% rounding.
Do you also round your BP readings up to numbers that are more pleasing to the eye?
 

IlDestriero

Ether Man
10+ Year Member
Nov 24, 2007
7,555
7,045
The ivory tower.
Status
Attending Physician
Do you also round your BP readings up to numbers that are more pleasing to the eye?
I have found that several of "our" "colleagues" are pretty creative with the old paper record. I like the electronic record much better. It actually reflects what happened. I used to give breaks, take over, etc and wonder if they write what they wanted the VS to be, or just filled out most of the record in advance.:mad: