If the pop-off valve is left closed during pre-oxygenation are you still delivering O2 to help preoxygenate the patient??
any thoughts?
any thoughts?
If the pop-off valve is left closed during pre-oxygenation are you still delivering O2 to help preoxygenate the patient??
any thoughts?
Assuming you have a perfect seal with your mask, the patient will be rebreathing if your fresh gas flow is equal to oxygen consumption (closed circuit anesthesia), if your gas flow is more than that (250cc/min) then for a short period of time he will rebreath until the bag becomes inflated like a big balloon and then there are 2 possibilities:I would think with a mask on for pre-O2 the patient would still be able to breathe in fresh O2 just not be able to exhale it out therefore...rebreathing??
Try this sometime during your morning machine check - close the pop-off, obstruct the circuit, and hit the flush valve. Keep doing it. The bag does not pop - you can inflate it to a ridiculous size! Also, you will notice that the pressure will not get much above the 40's, especially after it gets really big.
Do not recommend doing this with a patient however!
so are we saying then you can still adequately preoxygenate with the pop-off clsoed but make the patient hypercarbic from rebreathing??
What's really cool is going closed circuit. The bellows never reach the top of the canister nor the bottom. You can inject liquid anesthetic into the circuit and not crack the vaporizer. Only provide enough oxygen flow for O2 consumption and gas analysis.
Try this sometime during your morning machine check - close the pop-off, obstruct the circuit, and hit the flush valve. Keep doing it. The bag does not pop - you can inflate it to a ridiculous size! Also, you will notice that the pressure will not get much above the 40's, especially after it gets really big.
Do not recommend doing this with a patient however!
It is difficult to do closed circuit anesthesia with mechanical ventilation because the ventilator will not work properly at such low flows.
I do closed circuit always with spontaneous ventilation.
It's a nice technique but I think that the residents are not taught how to do it anymore.
it saves $$$ too doesnt it?
plus can you really do it with sevo at such low flows? yah yah i know the studies saying you need atleast 2 lpm were done on rats, but I thought it's still what's on the package insert?
actually, untrue. i put a peds FM with plastic seal still on and occluded circuit to prime as i went to get kid. (forgot pop off was closed). OR nurse comes into pre-op and said bag fell off machine. when i questioned what he meant, he couldn't tell me. we get back to the OR, put the kid on the table and i go to grab mask, only to notice the rez bag was shredded from pressure. crazy stuff. perhaps it was because it was a peds bag, but shredded none the less. so, these rez bags are not that infalliable.
It saves money
The time to reach a certain concentration is longer but you can increase the flow and increase the vaporizer's output until you reach the concentration you need then you go back to closed circuit and shut the vaporizer off, the vapor concentration will stay constant for a long time.2nd - The majority of ventilation is re-breathed gas. Fresh gas flow is only a small contribution to your overall ventilation. Therefore, you will have to dial in a higher concentration on your vaporizer than you want to have as an inspired concentration. Pay attention to your inspired and end tidal volatile agent levels and adjust your dial up and down accordingly. Don't forget to reduce your vaporizer settings when you increase fresh gas flow.
-pod
1st - Pay attention to your CO2 absorbent and your inspired CO2 level as you will exhaust your CO2 absorbent much more quickly than with "normal" flows.
If you have never done minimal flow anesthesia, let me give you three warnings.
1st - Pay attention to your CO2 absorbent and your inspired CO2 level as you will exhaust your CO2 absorbent much more quickly than with "normal" flows.
2nd - The majority of ventilation is re-breathed gas. Fresh gas flow is only a small contribution to your overall ventilation. Therefore, you will have to dial in a higher concentration on your vaporizer than you want to have as an inspired concentration. Pay attention to your inspired and end tidal volatile agent levels and adjust your dial up and down accordingly. Don't forget to reduce your vaporizer settings when you increase fresh gas flow.
3rd - Thoroughly question anyone who gives you a break as to their comfort with low flow anesthesia and abandon it for the duration of your break if they aren't completely on board. I have tried to teach several or our CRNAs, but I always come back from break to find things FUBARed. There are a couple of attendings who I would not trust either.
-pod
so are we effectively preoxygenating a patient even if the popoff is closed??
many a times ive seen residents forget to open the popoff valve when preoxygenating a patient with the mask on prior to going to sleep. obviously the patient is awake and therefore there will be a leak (patient messing with the mask) so patient will get uncomfortable sometimes but not always.
Another attending argued with me that the patient was still being oxygenated because the O2 was still coming thru...this did not make any sense to me...No ETCO2...the sat not increasing...i begged to differ