Some questions about vents...
With VC using my anesthesia vent, I have no way to control the actual inspiratory flow (not FGF) however I am able to adjust the I:E ratio. Is this because my machine just predicts the flow needed (dummy proof)?
For example if I set 600 mL for RR of 10 at I:E of 1:1, does that mean the vent just knows to deliver 600 mL in 3 sec (inspiratory flow at 200 mL/s).
Next question is even if I turn inspiratory pause completely off, my machine is still able to give me both peak and plateau pressures and looking at the Pressure/time tracing I can still see a peak followed by a quick drop to plateau then the expiratory drop. Does this mean the machine forces a quick inspiratory pause with each TV just so it can calculate the Plateau?
I know with our ICU vents with VC, the only way to adjust I:E ratios, the flows have to be adjusted (which then displays the I:E ratio). It also never displays a plateau pressure unless you manually force an inspiratory pause. It does however display the Pressure Mean which is a nice number to know for oxidation reasons, while our anesthesia vent doesn't.
With VC using my anesthesia vent, I have no way to control the actual inspiratory flow (not FGF) however I am able to adjust the I:E ratio. Is this because my machine just predicts the flow needed (dummy proof)?
For example if I set 600 mL for RR of 10 at I:E of 1:1, does that mean the vent just knows to deliver 600 mL in 3 sec (inspiratory flow at 200 mL/s).
Next question is even if I turn inspiratory pause completely off, my machine is still able to give me both peak and plateau pressures and looking at the Pressure/time tracing I can still see a peak followed by a quick drop to plateau then the expiratory drop. Does this mean the machine forces a quick inspiratory pause with each TV just so it can calculate the Plateau?
I know with our ICU vents with VC, the only way to adjust I:E ratios, the flows have to be adjusted (which then displays the I:E ratio). It also never displays a plateau pressure unless you manually force an inspiratory pause. It does however display the Pressure Mean which is a nice number to know for oxidation reasons, while our anesthesia vent doesn't.